COMMON PROBLEMS AND SOLUTIONS
1.Presentation of method validation data. Table 3.2 provides a quick over-
view of the validation data.
2.System suitability. During the robustness testing of method validation, criti-
cal method parameters such as mobile phase composition and column temperature
are varied to mimic the day-to-day variability. Therefore, the system suitability
results from these robustness experiments should reflect the expected range for the
system suitability results. As a result, system suitability results in these method
validationexperimentsareveryusefulindeterminingthesystemsuitability
acceptance criteria. This is a very effective approach since the required system
suitability results can be generated during method validation and no other special
study is required. However, these results reflect the expected performance of the
system, but not necessarily the minimum performance standard for acceptable
results. For example, the minimum resolution of the critical pair from method
validation may be 3.5; however, a resolution of 2.0 may still be acceptable as
longas theyare baselineresolvedandallotherchromatographicparameters
remain acceptable. |
|