Buffer. Ensure that the buffer (pK a ) is appropriate for the pH of the solution.
In general, to provide appropriate buffering capacity, the pH of the solution
should be within ± 1 pH unit of the pK a value of the buffer.
žIsocratic method. Whenever possible, use isocratic HPLC condition, as this
is affected less by the variation in flow rate, temperature, and dwell vol-
ume. If gradient HPLC conditions have to be used, a simple linear gradient
is preferred over multistep gradients. Complicated gradient conditions are
more susceptible to differences between HPLC instruments (e.g., flow rate,
dwell volume). |
|