žDilution error. Pipette a volume of more than 5 mL, and avoid using volu-
metric flasks of less than 25 mL.
žSonication.Theefficiencyofsonicationishighlyvariableanddepends
on various factors (e.g., condition of the sonication bath, level of water,
and position of flask in the sonication bath). Mechanical shaking is recom-
mended, instead, and is much more reproducible.
žMobile phase as sample solvent. If possible, always use mobile phase as the
sample solvent. This ensures the composition (e.g., percent organic, pH) of
sample solution matches that of mobile phase and reduces the chance of
any problem due to incompatibility of sample solvent and mobile phase.
Alternatively, always use sample solvent weaker than that of the mobile
phase to ensure that the chromatography is not deteriorated. For example,
in reversed-phase HPLC, use less organic solvent in the sample solvent than
in the mobile phase. |
|